2. Connect the DC power supply with EUT - Power 1. When ever possible connect the positive pin of the
dc source with L and the negative pin with PE. Only when a grounding of the EUT is no allowed connect
the negative pin with the neutral N of the TRANSIENT TRA2000IN6.
它建议用直流源正极接L端,用负极接PE端(这样才能提供线对地的12Ω阻抗),只有当EUT没有接地时,才把直流源负极接N端。
SURGE superimposing on dc
For this kind of test the dc voltage should be, if possible connected to L and PE on EUT-Power 1. Only for
this coupling path the coupling impedance is 10 Ω and 9 μF.
This is an advice from EMCP based on experience of customers, that protection devices has been
destroyed when SURGE test has been carried out with coupling impedance 18 μF and 2 Ohm between L
and N. In the real installation environment they never had a damage of the equal protection devices. In the
IEC 61000-4-5 no specific chapter is dealing with different d.c. sources. The only hints for Surge tests on
d.c. supply can be found in the single phase test set up examples.
这样的操作(正极接L端,用负极接PE端)是因为客户的经验:发生器有效阻抗选择2Ohm时,实际环境中能很好工作的保护元件会经常损坏,
所以选择12Ohm的阻抗,这样施加到DC电源线上的能量就小一点,EUT就不容易损坏。