5.1.2 Artificial mains network:人工电源网络
The artificial mains V-network is required to provide a defined impedance at high frequencies between the terminals of the equipment under test and reference ground, and also to isolate the test circuit from unwanted radio-frequency signals on the supply mains.
The artificial mains V-network 50 Ω/50 μH (or 50 Ω/50 μH + 5 Ω) as defined in Clause 4 of CISPR 16-1-2 shall be used.
To ensure that, at the frequency of measurement, the impedance of the mains does not materially affect the impedance of the artificial mains V-network, a suitable radio-frequency impedance shall be inserted between the artificial mains V-network and the supply mains.
This impedance will also reduce the effect of unwanted signals which occur on the supply mains (see also 5.3).
The connection between the artificial mains V-network and the measuring receiver shall be made by means of a coaxial cable with a characteristic impedance of 50 Ω. |